Title: High-pass filter having poor ECG transient response Post by gido on Mar 22nd, 2012, 12:22pm Hello, I have designed a fully differential passive high-pass filter for ECG and in running the transient response, I get the common mode to be decreasing and settling at a longer time. How best can i resolve this? ( make the common mode zero)..... Attached is the fully differential transient response. Thanks |
Title: Re: High-pass filter having poor ECG transient response Post by gido on Mar 22nd, 2012, 12:24pm Pls attached too is the schematic of the passive fully-differential high-pass filter. I subtract the voltages at the two outputs to get the response |
Title: Re: High-pass filter having poor ECG transient response Post by wave on Mar 22nd, 2012, 1:45pm With limited information, this seems fundamental. Single HPF have settling times and a differential step on a pair may have asymmetry. Is your input DC same as output CM (0V) ? Are your waveforms DC balanced? You might want to plot single ended inputs, with DC content. How to you plan to amplify this? Wave |
Title: Re: High-pass filter having poor ECG transient response Post by gido on Mar 22nd, 2012, 2:07pm wave wrote on Mar 22nd, 2012, 1:45pm:
yeah, my input DC is same as output CM(0V) I plotted single ended transient and the DC contents are same(0V). One thing is, the fully differential transient response for a sinusoidal input has a constant CM that is zero throughout. |